   I use Agilent 4352b(a Signal Analyze) and 8664A(a Signal generator).4352b connect  8664A through a GPIB.

  Now I want to connect 4352b through Keysight 82357B (a GPIBtoUSB converter) .

  I search out some GPIB port :

  GPIB0::16::INSTR

  GPIB0::17::0::IINSTR

 GPIB0::17::1::IINSTR

 GPIB0::17::13::IINSTR

  GPIB0::17::14::IINSTR

  GPIB0::17::15::IINSTR

  GPIB0::17::16::IINSTR

   but they all can't connect !

  why?

0 Kudos
Message 1 of 5
(4,233 Views)

Why do you want a second GPIB interface. One GPIB interface can handle up to 15 instruments but two should always work. Just use a GPIB cable to connect the two instruments and you are done.

greetings from the Netherlands
0 Kudos
Message 2 of 5
(4,227 Views)

thank you 

I think you may have some misunderstandings, I use 2 instruments instead of two interfaces/ports/GPIB .

Computer——Signal Analyze ——Signal generator ,They are connected to each other.

Signal Analyze ——Signal generator,Analyze must connect to Generator to work, and they're already connected(This is the case at the factory).

Now I want to connected Signal Analyze.

0 Kudos
Message 3 of 5
(4,217 Views)
Solution
Accepted by xiaochouyu

Is the Generator or the Analyzer set up as the GPIB master?  You can only have 1 master on a bus.  Typically, if there is a "master" GPIB connection on an instrument, there is a second GPIB interface in order to accept commands from a PC.



There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 4 of 5
(4,209 Views)

thank you

It seems that I can't connect to Analyzer. Because it's going to control Generator

0 Kudos