Instrument Control (GPIB, Serial, VISA, IVI)

cancel
Showing results for 
Search instead for 
Did you mean: 

How to determine the reference level crossing to be counted for measurement

Solved!
Go to solution

Measuring Reference-Level Crossings - NI

usan1_0-1741329069166.png

I don't understand why the next second-level crossing point after the first crosspoint isn't counted

I think the signal goes outside the window and then crosses the reference level

 

Could someone clarify this for me?

0 Kudos
Message 1 of 2
(143 Views)
Solution
Accepted by topic author usan1

Missing information is that the hysteresis is symmetrical around the reference level (two dotted lines above and below)

 

Count a point if,

  • fall down crossing upper threshold and reference
  • rises up crossing lower threshold and reference
  • switch between fall down or rise up logic only if the signal crossed the threshold

santo_13_0-1741630118290.png

The middle peak points is not counted because the signal did not cross the lower threshold after the first sample.

Santhosh