07-22-2018 03:25 AM
Good morning (at least here
),
I try to connect to a Programmable DC Power Supply - iDRC, DSP-060-025HD.
I connected the instrument to the Ethernet connection on my laptop directly.
When I open NI MAX i see -
(What does the warning sign mean?)
Trying to open VISA test panel I get -
And yet I am able to access the device via chrome browser -
And indeed the VISA address string does match whats displayed on MAX.
I was also able to successfully ping the device.
Labview visa open with the same VISA address string gives the same error as MAX.
Anyone has an idea about what the issue could be?
07-26-2018 10:06 AM
Hi!! Have you tried uninstalling and then installing back the drivers?
There's a similar forum posted a few years ago:
Hope this helps!!
07-29-2018 12:53 AM
Hi,
Thanks for your reply.
I managed to get over the problem eventually.
Two things caused my issues -
1. Seeing as I can communicate with the device via browser I didn't bother to set my IP as static at first.
2. According to instructions, I set up an IVI resource name on VISA which was not functioning properly.
when I used the regular VISA resource name AND set static IP things were working properly.