Hi,

 

I have recently used NI MAX and NI TestStand along with IVI drivers to get my devices working(using USB). It worked for TG5011 function generator and TDS1002B Oscilloscope successfully.

 

Following the same procedure: I started off with 2000 model  DMM from Keithley with GPIB.

I am facing trouble in finding my device on NI-MAX devices and interfaces.I am using a KUSB-488B GPIB to USB adapter to connect it.

I can interact with the device from VISA Interactive Control and KUSB Diagnostic tool and can see the device in Windows Device Manager and KI-Configuration Utility. Surprisingly I cannot see GPIB Board in NI-MAX.

 

Please help.

 

System Details: NI-VISA 5.2 installed on Windows 7, KUSB-488 driver installed in NI Command Compatibility mode.(only one ieee_32m.dll is present),Un Installed NI-488.

 

Thank you

Sahithi

 

 

0 Kudos
Message 1 of 7
(6,090 Views)

I think MAX only works with the NI VISA and NI GPIB drivers.  I'm not sure it works with the Keithley drivers and GPIB controller.


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 2 of 7
(6,080 Views)

I am using IVI drivers to get these working on NI TestStand. I just had a look at IVI website and wanted to know if installing a VISA Shared components would solve this issue???

 

Note: I have the Keithley 2000 IVI driver from National Instruments.

 

Please help.

 

Many Thanks

Sahithi

0 Kudos
Message 3 of 7
(6,074 Views)
The best thing would be to use an NI controller. You've wasted more time and money trying to get the Keithley to work than any small cost difference there is between the two.

If you don't switch, then make sure you enable tulip support in MAX and the VISA configuration.

I would also suggest you contact Keithley. I don't think this is actually a problem with the NI software.
0 Kudos
Message 4 of 7
(6,056 Views)

@Dennis_Knutson wrote:


I would also suggest you contact Keithley. I don't think this is actually a problem with the NI software.

Not only that, but if you are going to use IVI, I suggest you pay a visit to their website and read their Getting Started guides.

0 Kudos
Message 5 of 7
(6,053 Views)

Thank you all. Will contact Keithley for support.

 

Regards

0 Kudos
Message 6 of 7
(6,037 Views)
Solution
Accepted by topic author Sahiti2013

To help anyone who read this thread:

 

I have got the Keithley 2000 working from NI-MAX and NI-Test Stand Using IVI Driver(for Keithley 2000 from NI)and using KUSB-488B(GPIB to USB adapter from Keithley).

 

Installed Keithley I/O layer from keithley and I can see the device in Hardware assets. I then followed the procedure for IVI  driver program MAX and TestStand and IT WORKED !!!!!!

 

Please find attached ScreenShots for Visual Clarity.

 

Thanks and Regards,

Download All
0 Kudos
Message 7 of 7