Paralell talking on the GPIB bus
My system uses GPIB bus. I controll that using VISA libraries from LV.
One controller and ten devices connecet to the bus.
Can I have two session opened - one sesion per device - and receive (to the controller) the data from both of them simultanousely?
(5,404 Views)
You can only talk with one instrument at a time due to hardware reasons.  But the hardware can figure that out.  Just understand that they won't be EXACTLY in parallel.
(5,400 Views)
OK,
Is it a GPIB limitation? What happen if I use LAN instead GPIB?
@crossrulz wrote:
You can only talk with one instrument at a time due to hardware reasons. (...)
Is it an assumption of the GPIB bus? Could you refer me to any document which states this?
(5,395 Views)
the only parallel action on gpib is the GET this means Group Execute Trigger.
On LAN also only a message can be sent to one receiver, so no parallellism
(5,389 Views)
LAN has the same issue.  It is a matter of having only one bus that can be talked over at a time.  But the transmission is fast enough that you won't notice.
If you try to send a command through VISA to multiple instruments, it will be a race condition of which one will talk first.  99% of the time, it doesn't matter since the second one will talk as soon as the first is done.
(5,380 Views)
Please look at triggering if you are concerned at simultaneous acquisition. There is the software GET that Albert mentions and external hardware triggers.
(5,376 Views)