Good evening,

 

I am trying to make a thermocouple measurements with PXIe-2527 and DMM 4080 based on the accompanying example by NI.

My switch config - 

switch init.PNG

Scan list - "ch0:ch10->com0;"

My switch config

DMM init.PNG

Than what I do is basically - 

1. Set 10 thermocouple  channels

2. start scan 

3. measure with DMM

4. Repeat for all required thermocouple values

measure.PNG

 

for some reason I find that channels are skipped with the following pattern:

results.png

Notice the second row in the fist column the third row in the second channel...

 

Any one had similar issue?

Am i doing something wrong with my scan config?

 

Thanks,

Barukh

0 Kudos
Message 1 of 2
(2,986 Views)
Solution
Accepted by topic author Barukh_t

Such a stupid mistake...

 

If you scan 10 channels, your scan list should be 0:9! 

Well I guess sometimes talking about the problem helps you solve it...

 

 

Message 2 of 2