I am trying to setup a USB-RS485 device as per this knowledge base document:

https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019MPySAM&l=en-US

 

However all of the Advanced options under the serial com port are grayed out and changing the transceiver mode from 4 wire to 2 wire is not possible. How can I enable these?

Thanks.

0 Kudos
Message 1 of 4
(4,704 Views)

Hi ah.bic,

 

Are you using a serial card or your machine's built in serial ports? Have you confirmed that these ports are configured correctly (IDN command, loopback test, etc)? Finally, what version of the NI-Serial drivers are you currently using? You should update to the latest NI-Serial drivers if you haven't already.

 

I hope this helps!  

 

Andy C.

Applications Engineering

National Instruments 

0 Kudos
Message 2 of 4
(4,675 Views)
Solution
Accepted by ah.bic

On Windows Vista and later, these settings can only be modified by running MAX as Administrator. Just right-click on the icon, and choose "Run as Administrator", and everything should be available to you.

Message 3 of 4
(4,650 Views)

Andy, Jason,

 

The device is a National Instruments USB to RS485/422 interface cable (https://www.ni.com/en-us/shop/model/usb-485.html). Jason's solution of running NI MAX as admin provided access to the settings.

 

 

Many thanks.

 

Al