12-08-2021 10:57 PM
I connect my electrometer with TCPIP SOCKET and use VISA test panel input/output.
When I tested write, clear operation, could see clear message at Return Data.
However, when reading, it was confirmed that the following error message (VISA: (HEX 0XBFFF0015) Timeout expired before operation completed) appeared.
I don't know how to get rid of this message.
Please help.
Thank you.
Solved! Go to Solution.
12-09-2021 03:01 AM
Hello
Does the instruments used support IDN command?(*IDN?)
If it is not supported, there will be no response and it will time out.
If it is supported but times out, the termination character(CR or LF or CR+LF) may be incorrect.
12-09-2021 08:33 PM
Hello
Thank you for your advice
I checked my instrument's manual (Keithely 6517b). It supports IDN command(*IDN?).
Also I tried the termination character (CR[\r] or LF[\n] or CR+LF[\r\n]), the results are same as previous problem.
Is there other solutions of this problem?
Please let me know.
Thank you.
12-11-2021 02:51 AM
I looked at the 6517B datasheet, but it doesn't seem to have a LAN interface.
I don't think you can use SOCKET communication with this, but how do you connect your computer to the 6517B?
12-11-2021 03:55 AM
I used prologix gpib ethernet converter and connected 6517B to LAN.
Also I connected to SOCKET because I could not connect 6517B by auto LAN instrument.
12-12-2021
07:56 AM
- last edited on
09-05-2025
10:00 AM
by
Content Cleaner
Prologix controllers cannot be treated like regular GPIB controllers such as NI.
Please see the manual for details.
(http://prologix.biz/downloads/PrologixGpibEthernetManual.pdf)
[Main items that require individual operations]
-GPIB address specification
You need to specify the GPIB address to connect to for the Prologix controller. (++ addr command)
-Disable Auto read
The Prologix controller has a mode that automatically tries to read the response, but since some commands do not respond, it is recommended to disable it in the setting and send the read command individually. (++ auto 0 command)
・ Read command
If read is required, the ++ read command is required after sending the command. (There are three reading methods: until timeout / until EOI / until LF is received)
In addition, a separate license must be purchased to use NI-VISA without using NI hardware. (NI-VISA is not free)