VISA: (HEX 0XBFFF0015) Timeout expired before operation completed
I connect my electrometer with TCPIP SOCKET and use VISA test panel input/output.
When I tested write, clear operation, could see clear message at Return Data.
However, when reading, it was confirmed that the following error message (VISA: (HEX 0XBFFF0015) Timeout expired before operation completed) appeared.
I don't know how to get rid of this message.
Please help.
Thank you.
(3,713 Views)
Hello
Does the instruments used support IDN command?(*IDN?)
If it is not supported, there will be no response and it will time out.
If it is supported but times out, the termination character(CR or LF or CR+LF) may be incorrect.
(3,698 Views)
Hello
Thank you for your advice
I checked my instrument's manual (Keithely 6517b). It supports IDN command(*IDN?).
Also I tried the termination character (CR[\r] or LF[\n] or CR+LF[\r\n]), the results are same as previous problem.
Is there other solutions of this problem?
Please let me know.
Thank you.
(3,675 Views)
I looked at the 6517B datasheet, but it doesn't seem to have a LAN interface.
I don't think you can use SOCKET communication with this, but how do you connect your computer to the 6517B?
(3,666 Views)
I used prologix gpib ethernet converter and connected 6517B to LAN.
Also I connected to SOCKET because I could not connect 6517B by auto LAN instrument.
(3,663 Views)
Accepted by JinyongKim
Prologix controllers cannot be treated like regular GPIB controllers such as NI.
Please see the manual for details.
(http://prologix.biz/downloads/PrologixGpibEthernetManual.pdf)
[Main items that require individual operations]
-GPIB address specification
You need to specify the GPIB address to connect to for the Prologix controller. (++ addr command)
-Disable Auto read
The Prologix controller has a mode that automatically tries to read the response, but since some commands do not respond, it is recommended to disable it in the setting and send the read command individually. (++ auto 0 command)
・ Read command
If read is required, the ++ read command is required after sending the command. (There are three reading methods: until timeout / until EOI / until LF is received)
In addition, a separate license must be purchased to use NI-VISA without using NI hardware. (NI-VISA is not free)
(https://www.ni.com/en-us/shop/product/ni-visa.html)
(3,648 Views)