Xantrex XDL 35-5P
Has anyone been able to use the Zantrex Labview drivers to control the output on the AUX line?
I can use the drivers to control channels 1 & 2 but not the AUX line.
I have tried setting the channel name to both "3" & "AUX" but with no luck
Anyone any ideas
Thanks & regards,
Don1.
(4,908 Views)
OK Im a bit confused,
the XDL 35-5P is a single output 105W supply with the GPIB option.
Assuming you have a XDL 35-TP the only control over the AUX output is to turn it on or off.  OP30= turn off Output 3  and OP31 = turn on Output 3
the voltage supplied from the AUX output is selectable olny from the front panel
(4,898 Views)
My bad,  It is a XDL 35-5TP.
Thanks for the syntax, I am having some trouble execute it though.
I am using the NI-488.2 Communicator to test it out.
When I run write "*RST" or "*IDN?" or run a query such as "MODE?" these command work fine, however when I write "OP31", "OP11" or "OP21" the outputs never turn on.
Is there a specific mode I need to be in to run these commands.
Thanks & Regards,
Don1.
(4,891 Views)
Accepted by ShaneDonn
Solved.
it needs to be OP3 1 and not OP31
Thansk for all the help
(4,882 Views)
OOPS- sorry about that! Don't forget to mark the solution
(4,879 Views)
Thanks for posting the solution you found! Smiley Very Happy
Justin E
National Instruments R&D
(4,844 Views)
