8 m cables seem to work, but they technically violate the IEEE spec. If you are way off (I've seen 15 m cables), you can expect data corruption, and possibly slower performance (I would think the handshaking lines can take longer to settle?).Configuration Requirements
To achieve the high data transfer rate that the GPIB was designed for, you must limit the number of devices on the bus and the physical distance between devices. The following restrictions are typical:
For high-speed operation, the following restrictions apply:
- A maximum separation of 4 m between any two devices and an average separation of 2 m over the entire bus.
- A maximum total cable length of 20 m.
- A maximum of 15 devices connected to each bus, with at least two-thirds powered on.
- All devices in the system must be powered on.
- Cable lengths must be as short as possible with up to a maximum of 15 m of cable for each system.
- There must be at least one equivalent device load per meter of cable.
If you want to exceed these limitations, you can use a bus extender to increase the cable length or a bus expander to increase the number of device loads. You can order bus extenders and expanders from National Instruments.
‎03-09-2006 05:16 PM
‎03-12-2006 08:54 PM
‎03-13-2006 07:34 AM
‎10-08-2007 04:44 PM
‎10-08-2007 09:51 PM
‎10-09-2007 08:28 AM
‎10-09-2007
08:35 AM
- last edited on
‎03-18-2025
04:05 PM
by
Content Cleaner
Having 20 instruments would not be a problem as long as it is configured correctly. If you go this route you should consider the GPIB-120B