Hi,
You could try a couple of things.
In Measurement & Automation, Click on the Devices and Interface folder, then select the GPIB card (GPIB0). Then click on the Scan For Instruments. After a few minutes you should get a list Instrument in the right hand side of the screen of Measurement & Automation (MAX). Check if address 23 is there and if it is the Fluke 8842A.
If it there and not the fluke then you will need to change the addresses of one of the instruments to get rid of any conflict.
(You dont say if any other instruments are also connected).
If it not present and you are sure it is connected properly. If other instruments are addressed correctly, then try changing the bus cable to the fluke.
Re check the Interface settings on the instrument. Make sure its
not set for RS232 interface (if it has that capability).
Hope this gets you going.
Ray Farmer
Regards
Ray Farmer