hi, we have the ni-usb-gpib device and  using it with the ni communicater under max. when i search for instruments it finds my signal generator at gpib address 7, but under the id it says "error no leading command". i used the ni-communicator and enter in id?* but it still says "error no leading command".

 

how do i enter in ID?* with a leading command??can you show an example. do i have to config the "EOS" under ni-communicator(MAX)??

0 Kudos
Message 1 of 3
(3,631 Views)

Can you send a screenshot of the message? I've never heard of 'leading command'.

 

Your command of 'ID?*' is not standard. Are you sure that's correct. What is defined by IEEE-488.2 is '*IDN?'. Quite a difference.

0 Kudos
Message 2 of 3
(3,619 Views)
Solution
hi, thanks, we were able to get the unit talking for example we wanted to change the freq to 30hz of the signal generator and the command was FRQ:30;, but when i would enter this into the NI-communitor it would say "error no leading command", we tried different things finally the correct command was
"\n FRQ:30;". the signal generator is an old signal generator, but we were able to get all the commands working except for id?* the newer version of the same signal generator uses *IDN?
0 Kudos
Message 3 of 3