Agilent 34401A multimeter to PC with Labview driver via RS232
In order to read more signals, you will most likely need to reduce the resolution of your signal.  The higher the resolution/accuracy, the longer it takes to settle out.  The default settings on these meters is for 5-1/2 digits with a lower resolution option of 4-1/2 digits and a higher one of 6-1/2 digits.  Frequency and AC Voltage tend to take much longer to derive a value than Resistance or Vdc.
Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
Message 31 of 58
(2,714 Views)
Hello Gerd
In the manual of Agilent 34401a, I just checked, it is showing me to turn off Auto Zero to make the reading speed double.
I have tried finding that to make it zero, but still not able to see such command in menu/sub menu  option.
Can anyone tell me how to do Auto Zero , OFF in Agilent 34401A.
Message 32 of 58
(2,756 Views)
The way to turn AUTORANGE - Off is to select a particular range and resolution if desired.  To measure a 7-volt DC signal, The correct SCPI syntax would be
 MEAS:VOLT:DC? 10
which would set the voltage range to 10V and effectively turn off the autorange function.  By default, the resolution is set for 5-1/2 digits (10 PLC).  By adding the resolution to the end of the query it will change the PLC response time.
MEAS:VOLT:DC? 10,0.003
The AUTOZERO will follow:
Message 33 of 58
(2,715 Views)
Hi ashish,
Read page 59 of the manual!
The command is "ZERO:AUTO OFF"…
Message 34 of 58
(2,733 Views)
Missing Post
Message 35 of 58
(2,728 Views)
Hi all ,
Without any input to DMM Agilent 34401A , it is reading all reading noises etc and I am able to write that in a file using Labview.
But, WHEN I put input to dmm from vibrator  and read it through lab view it show 1073481730 error , saying voltage overload and on DMM screen it suddenly shows 0.VLD VDC, and my program stops.
How can I read my input measurement and what changes in setting required to correct it.
I am also attaching the snapshot of error occurs.
thanks and regards
Ashish
Message 36 of 58
(2,695 Views)
The Overload response is indicative of setting the range too low.  If you set the voltage range for 10V and have a 15V signal, you will get the overload response.
What are you trying to measure (Voltage, Frequency, Current,...)?
What is the peak value for this measurement type from the interfaced equipment (Vibrator)?
Message 37 of 58
(2,690 Views)
I am measuring VoltDC , the peak value is 10-12 volt. The range for this is 2 Volt to 12 volt DC.
Message 38 of 58
(2,687 Views)
The DC Voltage Range is 100mV, 1V, 10V, 100V, & 1000V range.  Make sure that the range is set to the 100V range to cover the max input of 12V with the best resolution.  If it was set to the 10V range, that may explain why you might see the overload condition.
Message 39 of 58
(2,680 Views)
Thanks alot, its working
Ashish
Message 40 of 58
(2,674 Views)