Fluctuation in torque measurement using NI 9237
I have prepared the LabVIEW FPGA program to measure torque meter reading. I am using NI 9237 for the torque meter. I have carried out calibration of torque meter by applying a range of weights and produced relation between voltage output and weight to convert voltage into Nm.
But when I started rotating the torque meter, observed that output voltage is fluctuating a lot and so torque value Nm. I acknowledge that I am using pretty much an old torque meter but it is working fine. I want to ensure that my code is correct for the torque meter. I am using the below code in LabVIEW FPGA to get voltage values from the torque meter (Mod6/AI0).
Could anyone suggest, is this the correct way or not? do I need to add anything else?.
I read in NI 9237 operating instruction that it has different filters like passband and stopband filter. I am not sure how to use it?
Thank you.
Any help will be appreciated.
Message 1 of 17
(2,855 Views)
My first question would be - how are you rotating it? How sure are you that the torque is "constant", for some given level of constant.
The torque sensor might be quite sensitive, and whilst hanging a weight is static and will nicely rest, motion might vary - maybe the motor is going a little faster, a little slower, oscillating around the target speed. How much variation in torque this produces can be significant, depending on the system.
Comparing with some other system doesn't necessarily help - if you connect e.g. a multimeter, probably it does some time-averaging before displaying a reading, which makes it look smoother. Of course, you could do this in LabVIEW too if you wanted, but you have to decide to do it, and then do it.
Is the mean value of your measurement (close to?) what you expect? How large is the variation?
What sort of a wiring setup do you have - how is the sensor connected to the cRIO, is there a slip ring, are you using large AC motors, does the wiring for the sensor have any protection from electrical noise, etc?
GCentral
Message 2 of 17
(2,849 Views)
As shown in below schematic diagram toque meter is installed between the machine (Roots blower, whose power I need to measure) and pully. It is connected with a motor using a belt drive. I am operating a motor using VFD, so I can run the machine at the required RPM. I am measuring RPM as well and it is pretty much stable.
Please refer below table of measured data at a constant speed. variation between maximum and minimum value is too high. So mean value is not representing the correct result.
I am using shielded cabled which has 5 wires, AI+,AI-,EX+,EX-,Extra (using as ground). these 5 wires are connected with RJ50 connection of NI 9237 using a connector available from NI. To avoid noise due to the power supply I have kept all power supply in a separate box.
Blower Speed  TM Voltage    Torque
    RPM        (V)      (Nm)
1188.068 0.000047 0.869199
1188.334 0.000072 1.677562
1188.275 0.000109 2.92506
1188.225 0.000105 2.787978
1188.38 0.000072 1.684008
1188.029 0.000086 2.17024
1188.634 0.000066 1.47826
1187.741 0.000154 4.427263
1188.117 0.000083 2.041834
1187.511 0.000099 2.596112
1188.159 0.00006 1.281561
1188.581 0.000108 2.890975
1188.052 0.000109 2.912542
1188.844 0.000137 3.856872
1187.794 0.000077 1.85505
1189.044 0.000111 2.977117
1188.728 0.000079 1.919006
1188.307 0.000145 4.119757
1187.791 0.000097 2.522861
1188.081 0.0001 2.635403
Message 3 of 17
(2,825 Views)
What is full scale torque for your torquemeter?
How much torque did you put on it when you were calibrating?
Your table of  data doesn't show any problems.  Your variations in voltage are sub-millivolt.  Electrical noise tends to be higher than that.  As a result, your torque fluctuates between ~ 1 to 4.5 Nm.   That is nothing in terms of torque.
Message 4 of 17
(2,791 Views)
Full scale of torque meter is 0-50Nm (5kgf m). it is giving output in mV/V. Below is the calibration table, I have applied weight to cover entire range.
Fluctuation from 1 to 4Nm is quite high for our measurement.
Nm (applied weight) Volt
-0.07 0.00002
0.54 0.00004
1.16 0.000059
1.75 0.000077
2.00 0.000084
2.25 0.000092
2.50 0.0001
2.75 0.000105
2.87 0.000107
3.00 0.000107
3.00 0.000109
3.49 0.000125
4.00 0.000142
5.00 0.000171
6.00 0.000203
7.00 0.000235
7.90 0.000262
8.00 0.000267
10.35 0.000341
12.80 0.000421
15.25 0.000501
17.71 0.00058
20.16 0.000661
22.61 0.000739
27.51 0.000898
32.42 0.001058
37.32 0.001216
42.22 0.001375
47.13 0.001534
Message 5 of 17
(2,778 Views)
Hi Patel,
You are getting just 1.5mV at full scale input?
Which supply voltage did you use?
Can you attach the manufacturer manual for that torque sensor?
Message 6 of 17
(2,771 Views)
I suspect some problem with excitation or amplification... 1.5mV at full scale seems likely what should be going into an amplifier before measurement.
I took a look through the FPGA properties for a 9237, and at the manual, but didn't see anything that jumped out. You can set the excitation voltage, which might need to be adjusted based on your sensor? Or you can provide an external value, in which case your power supply should be a) at the correct voltage and b) turned on! 🙂
2-4Nm on a 50Nm scale isn't great, but it should be ok (4/50->8% used). 1.5mV on your input scale seems like your problem.
The allowed input is 25mV/V, which with a 3.3V excitation (looks to be the default, you might have changed this?) would make full-scale 82.5mV - and then 1.5/82.5 -> just 1.8%.
GCentral
Message 7 of 17
(2,762 Views)
I am sorry about manufacturer manual, this is very old torque meter and we don't have manual available for specific model.
I am using 10V internal excitation from NI 9237 module.
Message 8 of 17
(2,754 Views)
Thanks, I am not using any kind of amplification, just directly connecting the torque meter output to NI 9237.
I am using 10V internal excitation from NI 9237.
Message 9 of 17
(2,753 Views)
How did you identify the wires? Is there no documentation at all available? Or perhaps a part number for the sensor? The manufacturer's name? The manual doesn't necessarily have to be for the specific model (although of course that would be ideal) - sometimes a manual covers a range of similar products?
GCentral
Message 10 of 17
(2,750 Views)