Hello RVH,
To remedy your problem, you can download the latest version of NI-SCOPE. The file is 854 MB so it will take some time. If you do not have the time to dedicate to downloading and installing the latest driver, you can simply copy the appropriate driver-associated files from the LabVIEW 7.0 directory to the LabVIEW 8.0 directory.
Best wishes,
Wallace F.
06-17-2008 06:41 PM
Thanks for this advice. We decided to download the Ni-Scope update. After this, the program was able to find all the necessary vi's itself and it could be run. Unfortunately, the PCI-5124 card still did not acquire the data. We ran a self-test on the card and got the following error message:
"Error -200313 occurred at Self Test
Possible Reason(s):
Measurements: Hardware clocking error occurred.
If you are using an external sample clock or an external reference clock, make sure it is connected and within the jitter and voltage specifications. Also, verify that the rate of the external clock matches the specified clock rate. If you are generating your clock internally, please contact National Instruments Technical Support."
We are not using an external clock.
06-18-2008 12:53 PM
Hello RVH,
Regarding Error -200313, I recommend resetting the PCI-5124 in Measurement & Automation Explorer and then perform another self test. If the device fails the self test again with the same error, you should call 866-275-6964 to speak with National Instruments technical support.
Best wishes,
Wallace F.