Hello folks,
I'm asking this question before I any post code or software versions etc. to see if there is a simple answer.
I'm using a PCMCIA 6036E card to read one analog input channel (using DAQmx...i.e. create task, create channel, set timing (continuous samples), start task, read in a software timed while loop (as fast as possible)). Regardless of how I set my sample rate, number of samples per channel (i.e. buffer size), or numer of samples to read, it looks like I can only get multiples of 512 samples.
The following lists some sample freq (Fs), the requested "number of samples to read", and the actual number of samples read:
Fs numberOfSamplesRequested numberOfSamplesActuallyRead
200Hz 20 512
5000Hz 1000 1024
5000Hz 2000 2048
QUESTIONS:
1. Is this "normal" behavior for a PCMCIA 6036E?
2. If not, has anyone who may have seen this issue determined the cause and how to "fix" it?
Best Regards,
Chris
Solved! Go to Solution.
12-18-2009 07:55 AM
chassan wrote:Hello folks,
QUESTIONS:
1. Is this "normal" behavior for a PCMCIA 6036E?
Sort of.
2. If not, has anyone who may have seen this issue determined the cause and how to "fix" it?
Best Regards,
Chris
On PCI systems DMA is used to transfer
the data, and daqmx get notified when the number of samples are
acquired. Now on PC-card that doesn't work, and the data is transfered
when the on-board buffer is full (after 2 kbyte of data) at 2 channels
10 Hz this can take a while.
There is a work around, there is a property where you can set the transfer mechanism (I don't have daqmx on this PC), or the polling mechanism of daqmx. Try these.
Found a KB document.
Ton
12-18-2009 09:12 AM
Hi TCPlomp,
Thanks for the great link; it explains exactly the situation that I'm experiencing!
I haven't tried the work-around yet, but at least I now understand why I'm seeing the behavior (i.e. it's not something in my code).
Best Regards,