10-29-2008 12:27 PM
10-29-2008 12:36 PM
Have you tried running that Basic code to see if it works? (Though I'm not sure if newer versions of Windows still supply Basica or Qbasic).
Are you sure you have the right cable? Null modem vs. straight through.
Have you tried communicating in Hyperterminal yet like Dennis mentioned?
10-29-2008 01:00 PM
10-29-2008 01:55 PM
RTS and DTR? must not be asserted or the meter will not send data. Some details here on pg6:
http://www.utsc.utoronto.ca/~quick/PSCB01H3S/Manual/ApA.pdf
and on pg11 here:
http://www.utsc.utoronto.ca/~quick/PSCB01H3S/Manual/Ch05.pdf
As already suggested, verify your cabling and start with Hyperterminal.
-AK2DM
10-29-2008 01:55 PM
The Basic programm is working
I have checked the serial port with Hyperterminal, I type "D", but there is nothing feed back on the screen.
10-29-2008 03:12 PM - edited 10-29-2008 03:14 PM
AnalogKid2DigitalMan wrote:RTS and DTR? must not be asserted or the meter will not send data. Some details here on pg6:
http://www.utsc.utoronto.ca/~quick/PSCB01H3S/Manual/ApA.pdf
and on pg11 here:
http://www.utsc.utoronto.ca/~quick/PSCB01H3S/Manual/Ch05.pdf
As already suggested, verify your cabling and start with Hyperterminal.
-AK2DM
I found the first item on page 8 of the ApA.pdf. That seems backwards to me for a normal serial protocol that the lines should be unasserted. ![]()
![]()
But that explains why it won't work in Hyperterminal. When you open the port in hypertermina DTR and RTS go high. I don't know if there is any way to force them to be low from within Hypterminal.
Within LabVIEW, those lines can be set to low.

(Also double check your baud rates.)
10-29-2008 03:31 PM
RavensFan:
I agree, probably the best best would be to leave them unwired (physically in the real world, not virtually in LabVIEW), just use the RXD, TXD, and GND wires since there is no flow control protocol.
Sorry I did not preview your attachment.
-AK2DM
10-29-2008 03:56 PM
The problem has been solved. Thank everyone
AnalogKid2DigitalMan gave me two useful PDF.
The problem is "The instrument does not support handshaking, but the RTS line must be unasserted or else the DMM will not send.6"
When I add a property Node and set RTS state = 0 ,then it is working.
01-20-2009 05:46 AM
01-20-2009 06:13 AM
Where do I have to add the property node? Here: