Hi!.

I've created this VI that perform a spectral analysis of a simple sinusoidal signal of wich I can change the frequency in a range from 100 to 1000 Hz.

I've noted that if I choose a frequency's value like 340,350,..., 460... and so on, I see in the spectral graph exactly the expected signal's amplitude value, in this example: 4

uno.pnguno.png

Instead, If I set a value like 341, 342, 343,... and so on until the next "round" value I don't see the same exact amplitude value (4) but a lowest value and the graph is more scattered around the max value:

due.pngdue.png

 

Any idea?

Thanks in advance

0 Kudos
Message 1 of 6
(4,020 Views)
Solution
Accepted by topic author mforbiz

You take 1000 samples which is 100 ms of data; your frequency resolution is 10 Hz. When your frequency doesn’t fall exactly on a 10 Hz bin, it is spread out over multiple bins depending on the window.

 

mcduff

Message 2 of 6
(3,995 Views)

McDuff is right, add another 0 to your sample rate and you'll get perfect 1-Hz matching again.

0 Kudos
Message 3 of 6
(3,983 Views)

You only get a single line if the sample contain an integer number of cycles.

If you don't have an integer number of cycles in the signal, you get spectral leakage.

 

 

0 Kudos
Message 4 of 6
(3,946 Views)

Hi mcduff.

Sorry for delay.

You're right.

I'll try with windowing

Thanks

0 Kudos
Message 5 of 6
(3,885 Views)

Hi Bert.

I think you intend to add another 0 to number of sample.#Sample.png#Sample.png

0 Kudos
Message 6 of 6
(3,884 Views)