Spectrum analysis, strange behaviour
Hi!.
I've created this VI that perform a spectral analysis of a simple sinusoidal signal of wich I can change the frequency in a range from 100 to 1000 Hz.
I've noted that if I choose a frequency's value like 340,350,..., 460... and so on, I see in the spectral graph exactly the expected signal's amplitude value, in this example: 4
Instead, If I set a value like 341, 342, 343,... and so on until the next "round" value I don't see the same exact amplitude value (4) but a lowest value and the graph is more scattered around the max value:
Any idea?
Thanks in advance
(3,981 Views)
Accepted by mforbiz
You take 1000 samples which is 100 ms of data; your frequency resolution is 10 Hz. When your frequency doesn’t fall exactly on a 10 Hz bin, it is spread out over multiple bins depending on the window.
mcduff
(3,956 Views)
McDuff is right, add another 0 to your sample rate and you'll get perfect 1-Hz matching again.
(3,944 Views)
You only get a single line if the sample contain an integer number of cycles.
If you don't have an integer number of cycles in the signal, you get spectral leakage.
(3,907 Views)
Hi mcduff.
Sorry for delay.
You're right.
I'll try with windowing
(3,846 Views)
Hi Bert.
I think you intend to add another 0 to number of sample.#Sample.png#Sample.png
(3,845 Views)