I've been trying to solve this problem for a while now without any luck. Any solution suggestions are welcome.
I am using a PXI 6232 with LabView 8.5.1 to input several analog signals from my sensors. Based on the signals a PWM signal is generated and output using PXI 6713.
Some of the analog input signals have spikes in them, that continuously occur during the runs. I looked at the same signals on an oscilloscope - no spikes. I change the setup of my hardware and the spikes still occur at the same spots. It seems like the program performs resets causing measurement errors.
I attached the VI and a JPEG of two measured inputs.
Thanks in advance
Regards
Vadim
Solved! Go to Solution.
04-08-2010 02:17 AM - edited 04-08-2010 02:20 AM
I first was confused by your timescale 😉 but it seems that these spices occur every 20ms (not s) which point to a 50Hz line noise due to switched power converters (or an uncompensated diode bridges 😮 )
Another hint was your scope measurement. While using the scope you opened a groundloop so the spikes due to dI/dt through the groundloop found another way around.
So I'm pretty sure that it is not the DAQ (in this case) it's your setup.
Provide a clean (low R AND low L) power path (keep them close and twist them slightly if possible) , add some filter to get down the dI/dt , identify the ground loops . (Use your scope with a little coil as a sensor at the input and catch some magnetic fields can open eyes)
THEN to clean the last spikes (out of the DAQ) to get the last ppm use chokes
04-09-2010 02:56 PM
Thanks Henrik,
I really hoped it was the DAQ, cause I've been hunting the ground loops for a while now - without any luck. Got myself different sensors/transducers, which changed the system' grounding connection slightly - no spikes (at leas nothing significant.)
Thanks again.
Regards
Vadim