CTR Z not working?

Hi,
I'm reading an encoder with a PCI 6221, using the signals CTR 0 A, CTR0 B, and CTR 0 Z.
It seems that the reference (CTR 0 Z) doesn't work some times. I have measure the signal wired to CTR 0 Z with an oscilloscope and the signal is Ok (I mean the voltage level), but I don't know if the signal durations (in miliseconds) is "enough". Which is the minimum pulse width for this signal?
Thank you in advance,
0 Kudos
Message 1 of 2
(2,738 Views)

The Z pulse must be in the high state simultaneous with the A,B signals being in their configured state. There are 4 possible High-Low permutations to pick from for A,B and I don't recall the default offhand (H-H maybe?).   So it's not exactly a minimum pulse time that's needed, it's more of a pulse sync which may not be possible for you to do without some interface circuitry.  

This restriction works fine with industrial encoders that often generate a Z pulse that is synced to A,B and lasts either 1 A,B state or 1 quadrature cycle of 4 states.  It doesn't work well when you'd like to reset a count value based on some kind of asynchronous external signal (home sensor, etc.).  Another note - the behavior is *level*-sensitive, not edge-sensitive.  If you hard-wire Z to a 5V High, the count will simply cycle 0,1,2,3,0,1,2,3,0,1...

There really isn't any simple workaround that I know of.  You might want to visit the Product Suggestion Center to add your voice.  Personally, I think counters should be allowed to hw-reset their count based on an asynchronous edge, even in modes other than position measurement.

-Kevin P.