07-15-2008 03:37 PM
Best regards
07-16-2008 09:08 AM
Javier,
cFP should work great for this application. It looks like you found the right modules for the job. I do recommend 4-20mA pressure sensors when possible to help with noise immunity in your system.
07-16-2008 09:39 AM
Hi, I work at MWM International (Brazil) and we actually use Compact Field Point to do the automation. (Two backplanes communicating with a PC from ethernet bus).
It work very well. The only problem I think about the cFP in this case is the fact that it haven't support to CAN bus.
But I recommend it.
07-16-2008 10:52 AM
07-16-2008 12:02 PM
We are monitoring:
12 pressures (4 to 20mA transducer)
15 temperatures (10 tcK and 5 RTD)
5 frequency channels (engine speed, mass air flow, and others)
6 analog inputs (throttle position, and others)
14 digital inputs (emergency button, test bench door open, dynamometer failure, and others)
8 analog outputs (several controllers setpoints)
28 digital outputs
All channels divided in two backplanes
07-16-2008 12:20 PM
Javier,
I work in a lab that tests submersible motors and pumps. Here are a few of the measurements I regularly make with cFP.
Pressure with 4-20mA transducers.
Temperature with thermocouples and PT100s. Isolated temperature measurments using Analog Devices 5B series of modules and cFP-AI-110 module. 5B modules provide channel to channel isolation. This allows for electrically 'hot' thermocouples that are connected directly to motor windings.
Fluid flow using turbine and mag flowmeters with 4-20mA outputs.
I am also using the serial ports on my cFP-2120 controllers to interface with many other instruments in the lab.
07-16-2008 11:04 PM
07-17-2008 07:07 AM
javier,
As an example, when testing a pump/motor, I store volts, amps, watts, Hz, GPM and PSI for each test point. While the test is running I hold the test data in an array. This information is written directly into an Excel template at the end of the test. Also store all the test parameters entered by the operator into the Excel file. As the test is running, I graph the raw PSI vs GPM so the operator can monitor the test.
Not sure how the USB-DAQ device would hold up in that environment. I would consider using either the cFP or the cRIO.
07-17-2008 12:09 PM
07-18-2008 10:32 AM