Hmm. the message I get is Error # 200077 "Requested value is not a supported value for this property." when I set the gain to 250. Same error I got when I put the gain all the way down to 10.
Did you use the simulated device or the real hardware? Also, is there a way to adjust the input voltage range on the DAQ communicator board? I am using the PXI-6070E and have not
seen a way to adjust this when doing a thermocouple measurement.
Any ideas?
11-14-2006
01:48 PM
- last edited on
08-06-2025
08:56 AM
by
Content Cleaner
Hi Ruz,
I tested with the simulated devices, but there should be no difference in the behavior when selecting configurations in software. Do you get differing behavior if testing this with simulated devices? Have you verified that the jumper settings on your SCXI-1121 match those that you set in MAX? Please refer to page 2-8 of the SCXI-1121 User Manual here for more detailed information on the proper input channel jumper settings.
Because the thermocouple acquisition task is set up using the analog input channel on the SCXI-1121, all allowable settings in the task are based off of this device. Manually changing properties on the PXI-6070E is only possible if it is the device specifically set up for the task.
Regards,
Andrew W
National Instruments