'Error code =-10842'+'Calibration'+'PCI6111E'-=>'Hardware failure' or just need a 'Calibration'?//Happy Xmas!
Readings and outputs of PCI-6111E differ from the expected value sharply in order of magnitude(Please refer to
the question "Anybody confused by PCI-6111E or PCI-6110?" posted on 12/14/2002).So I want to cope with this by
doing a calibration:
Description: PCI-6111E
Calibration Date: 23-OCT-99
Calibration Interval: 12 Months
Calibration Due : 23-OCT-00
Warm up time: >30min
System Resource : IRQ9(Win2K),IRQ5(Win98)
Calibration routine runs under (Win98,VC6.0,NI-DAQ 6.9.3)
(adopted from the NI-DAQ example programme C:\Program Files\National Instruments\NI-DAQ\Examples\VisualC\Calib\Calibrate_Eseries.C)
status=Calibrate_E_Series(deviceNumber,calOP,setOfCalConst,calRefVolts)
iRetVal = NIDAQErrorHandler(iStatus, "Calibrate_E_Series", iIgnoreWarning);
===========Step1:No external connections to the board===========
=====Version1:=====
status=Calibrate_E_Series(deviceNumber,ND_SET_DEFAULT_LOAD_AREA,ND_FACTORY_EEPROM_AREA,0.0)
// status is 0
// iRetVal is 0
Pogram Output:
Measuring offset at AI channel 1 before calibration.
NOTE: Make sure the inputs are grounded.
The offset at AI channel 1 before calibration is -11.690000 ADC units binary
...Starting E-series calibration...this may take a while.
...E-series calibration completed!
Measuring offset at AI channel 1 AFTER calibration.
The offset at AI channel 1 AFTER calibration is -11.600000 ADC units binary.
Press any key to continue
=====Version2:=====
status=Calibrate_E_Series(deviceNumber,ND_SELF_CALIBRATE,ND_NI_DAQ_SW_AREA,0.0)
// status is -10842,and the program prompts to choose continue or not.
// iRetVal is 6
Pogram Output:
The offset at AI channel 1 before calibration is -11.850000 ADC units binary.
An error occurred during calibration. Please check errors.
===========Step2:ACH0+,ACH0-,ACH1+,ACH1-,AIGND are all short connected together===========
=====Version1:=====
// status is 0
// iRetVal is 0
Pogram Output:
The offset at AI channel 1 before calibration is -32.120000 ADC units binary.
The offset at AI channel 1 AFTER calibration is -32.030000 ADC units binary.
=====Version2:=====
// iRetVal is 6
Pogram Output:
The offset at AI channel 1 before calibration is -32.470000 ADC units binary.
I have searched and read several ideas on"Calibration" and "-10842" in the Discussion Forums and I wanna know
1.Is this card defective,sure???
2.Anybody can tell which part of the board is most probably defective?
3.What else?
Thanks a lot!
Happy Xmas!
hwlee@sina.com
(2,600 Views)
Hello;
One possible cause of that problem is that particular board is being calibrated a few times and it now needs an internal calibration. That type of calibration can be performed by NI only.
Anyway, even if that is not the case, since you keep getting that error, by using a example calibration code, your device, is most likely damaged, and the best way to go about that is to contact NI to have it repaired.
Regards
Filipe A.
(2,600 Views)