11-11-2009 10:36 PM
11-12-2009 03:42 PM - edited 11-12-2009 03:44 PM
Hi Oleks,
It sounds like you are trying to do a system level calibration. Typically, we would recommend calibrating each component seperately. The gain and offset error are specified for calibrated hardware and and won't be able to be compensated for. To calibrate your system together, you will have to have a strain guage with a known strain and zero out the errors associated with that particular strain guage.
11-12-2009 11:47 PM
Hi Jim,
Thank you for the response. I am not really trying to perform system level calibration. I should of being more clear about my problem. When all components are connected together, DAQ board, SCC units and a load cell, the only tool I was able to use to "zero out" signal from the load cell channels, was the resistors on SG24 units. However, even at "zero" level, I do have a noise, for instance Fx,Fy is ~+/-0.2N, while Mz ~ +/-0.002N with no load. Ideally I would like to get rid of that noise as much as possible, as it does not allow me to measure forces, say less then 2N if I want to have an error ~ 10% for Fx,Fy axis. So, what I am trying to figure out is if the part of that noise is actually come from gain and offset errors, and whether I can "software compensate" it.
If I understood you correctly, the only way to do that is by working on the loadcell hardware level. Is there an alternative?
11-16-2009 02:01 PM
Hi Oleks,
If you have a constant offset, you can easily zero this out in software. This Community Example shows one way you could do this. If you have noise on the system, Field Wiring and Noise Considerations for Analog Signals is a good resource for ways to reduce noise. You aren't going to be able to compensate for noise unless you are seeing a constant error - compensation can only add a constant offset, which will make no difference for noise. If you believe that the noise that you are seeing is out of spec, please let me know.