Is NI-9230 a good choice for PCB 393B12 accelerometer?

Solved!
Go to solution

Hello Community,

 

I want to measure floor vibration and the maximum expected frequency is 1000 Hz. 

Is NI-9230 a good choice for PCB 393B12 accelerometer? Do I need PCB signal conditioner or I can just connect them?

 

NI 9230 BNC, 3-Ch, +/-30 V, 12.8 kS/s/ch, 24-Bit, IEPE AI Module

 

Is the sampling rate of 12.8kS/s/ch seems enough for maximum 1k Hz data?

 

Thanks

0 Kudos
Message 1 of 5
(1,369 Views)
Solution
Accepted by topic author ShahabJHU

In general, 9230 is a good choice for vibration measurements (due to the 24-bit Delta Sigma ADC and excellent timebase)

 

Next, general rule of thumb, 10x the max signal frequency gives a very good sampled signal. But given this is vibration measurement, 10x sampling rate is plenty enough. So, yes, 12.8kS/s rate is great to capture 1kHz signal.

 

If you're looking to measure any kind of harmonics, the max harmonic is also limited by your sampling rate, in this case, a max 6.4kHz (Nyquist).

 

Refer to this article to see if your sensor signal is compatible with 9230's IEPE and signal range - https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000P8uwSAC&l=en-US

 

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
Message 2 of 5
(1,327 Views)
Solution
Accepted by topic author ShahabJHU

We use 353B31 accelerometers (similar sensitivity) for noise measurements on optical tables (or other vibration sensistive stuff) and  add a Kistler signal conditioner if we really want to digg in the noise and can't place big & heavy seismometers.

But by looking at the specks and if you don't want to go down or below 1 Hz* without chain calibration this combination tooks fine IMHO.

Edit: *)To measure mechanical transferfunctions in magnitude and phase with low uncertainty 😉

 

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


Message 3 of 5
(1,322 Views)

Thanks for the great help!

 

Is NI-9231 a better option for me? Because we need more than 4 channel. Could you please compare it vs 2 x NI-9234?

 

I noticed that my sensor output is±0.5g x 10000 mv/g=±5V. If I use NI-9230 with ±30 V input range, do I get lower resolution rather than using NI-9231 with ±5V input range? It seems NI-9231 is even faster 51.2 kS/s per channel update rate and 24-bit resolution. Indeed more expensive but it seems worth it!

 

Thank you,

 

0 Kudos
Message 4 of 5
(1,263 Views)

@ShahabJHU wrote:

Thanks for the great help!

 

Is NI-9231 a better option for me? Because we need more than 4 channel. Could you please compare it vs 2 x NI-9234?

 

I noticed that my sensor output is±0.5g x 10000 mv/g=±5V. If I use NI-9230 with ±30 V input range, do I get lower resolution rather than using NI-9231 with ±5V input range? It seems NI-9231 is even faster 51.2 kS/s per channel update rate and 24-bit resolution. Indeed more expensive but it seems worth it!

 

Thank you,

 


I would go with the 8 CH , but if you need to recable a lot, the extra space migth be more comfortable...

Concerning sync timing all in one looks better.

the resolution isn't your problem, the noise is of more concern.  the 9231 seems better by spec.

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 5 of 5