11-14-2012 09:09 AM
I have a NI-9211 that I would like to calibrate and adjust. But my Calibration Executive (version 3.3.2) only shows a verify step. Is there an adjustment step for this product in the newer versions?
11-15-2012 09:55 AM
Hi Saipanman,
Calibration Executive is only able to verify the operation of the 9211 (and various C seres modules). If the device shows to be out of spec, then it needs to be returned to NI to have the calibration constants adjusted. I checked in Calibration Executive 3.5 and this is still the case.
Regards,
Peter W.