Multifunction DAQ
NI 9234 Data Acquisition using a Kistler 8704B500
this may be a stupid question but I could not find any answers regarding this question.
I have got a Kistler 8704B500 sensor (http://www.intertechnology.com/Kistler/pdfs/Accelerometer_Model_8702B500_8704B500.pdf) and a C-Series Module NI 9234. Can I use them together?
I'm a bit worried because of the required current seems to be 4mA. Could you please clarfiy this for me? If not I can use different accelerometers but I first wanted to check!
Moritz
(3,106 Views)
Hey Mortiz
I've taken a look at the spec for that sensor. Is it IEPE type do you know? It doesn't specifically say.
Either way - if it requires a 4mA excitation current (which the specification states), then it's not going to work with the 2mA supplied by the 9234.
I'd pick something with a lower excitation current (or supply separately, i.e. from a current supply module)
Kind Regards
Chris | Applications Engineer NIUK
(3,089 Views)
