11-14-2017 10:09 AM
Hi,
I am using an NI USB-6009 device to control a pressure control valve through the NI board's analog output terminals. The user manual (http://www.ni.com/pdf/manuals/371303n.pdf page 17) clearly states that the board can generate voltage outputs from these two ports (AO0 and AO1) of 0 V to 5 V. However, when I am setting up and testing the DAQ connection in LabVIEW and I put a voltmeter to the analog output, the voltage on the meter reads between 0 V and 11.2 V. Any idea what is going on here?
Thank you for your help,
Richie
Solved! Go to Solution.
11-14-2017 10:59 AM
How is everything connected and terminated?
How much current are your valves drawing?
11-15-2017 06:21 PM
Hi altenbach,
Found out from the lab manager that this one was modified from manufacturer specs. So this is solved.
Thank you for your help though!
11-16-2017 08:04 AM
A DAQ HAQ!
-AK2DM