02-22-2009 07:02 AM - edited 02-22-2009 07:04 AM
02-23-2009
06:22 PM
- last edited on
07-23-2025
12:43 PM
by
Content Cleaner
I would recommend to review the document below because it discussions reasons as to why you may be seeing signal distortion and proper connection methods. It looks like you have a resistor to ground from the negative terminal but it is recommended to have a resistor to ground from the positive terminal also (with a value of 10kohms up to 100kHz).
Field Wiring and Noise Considerations for Analog Signals
02-23-2009 10:48 PM
02-24-2009
03:31 PM
- last edited on
07-23-2025
12:43 PM
by
Content Cleaner
Were you able to see if changing the sampling rate of your Daq device per Chris' advice on the other post that you have?
"Perhaps it's not really 4kHz noise, but something that aliases to 4kHz. The 9215 has a bandwidth beyond 400kHz. (http://www.ni.com/pdf/manuals/373779e.pdf) If there's 404kHz noise on the signal, for example, and you sample at 100kHz, or 50kHz, or 20kHz,... the 404kHz noise will appear to be at 4kHz. Try sampling around 99kHz instead of 100kHz. If the noise no longer appears to be at 4kHz, then it is aliased from a higher frequency. What is the bandwidth of the Dataq device?"