01-10-2013 09:28 AM
Hallo,
I have a problem about the PXI-2576 Switch.
When I open the MA-Explorer, I can see the Device PXI 2576, and also do the Selftest and Reset, they are OK, no problem.
But when I click the Button "Testpanel", there is always a error message "Error -200055 occurred at DAQmx Switch Get Relay Position (Multiple).vi:1" Details in the Photo.
Is there someone know, what should I do?
Thanks a lot!
Solved! Go to Solution.
01-11-2013 03:09 AM
Hi,
you can simply get rid of that error by removing the special characters in our devicename.
Below you will find i a screenshot where i changed the device name to e.g. PXI2576Multiplexer. It works.
I also tried it with the name that you used. -> it causes the error. -200055
i hope that info solved the problem also on your machine.
Andreas K
NI Germany
01-11-2013 03:48 AM
Hallo Andreas K,
Thank you very much for your reply!
I have change the devicename, in the MA Explorer, it works, no error message! Thanks!
11-27-2013 06:22 AM
Hey AndreasK,
Don't suppose you found or logged a CAR on this issue? I've just had a customer with the same problem and it seems like it's something worth CARing.
Cheers,
Chris