PXI-2576 Error Code -200055 by Measurement & Automation Explorer

Solved!
Go to solution

Hallo,

I have a problem about the PXI-2576 Switch.

 

When I open the MA-Explorer, I can see the Device PXI 2576, and also do the Selftest and Reset, they are OK, no problem.

 

But when I click the Button "Testpanel", there is always a error message "Error -200055 occurred at DAQmx Switch Get Relay Position (Multiple).vi:1"  Details in the Photo.

 

Is there someone know, what should I do? 

 

Thanks a lot!

0 Kudos
Message 1 of 4
(3,921 Views)
Solution
Accepted by Cabio

Hi, 

 

you can simply get rid of that error by removing the special characters in our devicename. 

 

Below you will find i a screenshot where i changed the device name to e.g. PXI2576Multiplexer. It works.

I also tried it with the name that you used. -> it causes the error. -200055

 

i hope that info solved the problem also on your machine. 

 

Andreas K

NI Germany

0 Kudos
Message 2 of 4
(3,911 Views)

Hallo Andreas K,

Thank you very much for your reply!

 

I have change the devicename, in the MA Explorer, it works, no error message! Thanks!

0 Kudos
Message 3 of 4
(3,904 Views)

Hey AndreasK,

 

Don't suppose you found or logged a CAR on this issue? I've just had a customer with the same problem and it seems like it's something worth CARing.

 

Cheers,

 

Chris

National Instruments - Tech Support