Quantization Error for NI AT-MIO-16E-2
Hi
I wanted to confirm the value of quantization error for the 12-bit NI AT-MIO-16E-2 DAQ. I can take values between -10 to 10 Volts. So the quanitization error would be:
= +/- 0.5*(20/(2^12)) = +/-2.44e-3
Is this correct? This seems like a high amount of error when voltages are below 0.1V
Also, I want to know if the above DAQ card comes with an anti-aliasing filter, depending on the sampling frequency.
Thank you
Karthik
(3,491 Views)
Hi Karthik,
If you choose your measurement range to be +/-10V, the smallest change that can be detected is 20/2^12 = 4.8mV. So the maximum quantization error would be +/-4.8mV. If you're looking to increase the accuracy, you can change the measurement range to be the smallest possible for your requirements.
There's no anti-aliasing filter built in to the DAQ card, but you can easily implement them on a terminal block (analog) such as the SCB-68 or in software (e.g. Low Pass Filter in LabVIEW).
Analog Input Low Pass Filtering with an SCB-68 Terminal Block:
http://digital.ni.com/public.nsf/websearch/5D82A2990FB1D21386256EEE00018FF0?OpenDocument
Hope this helps!
Lesley Y.
Applications Engineering
(3,483 Views)
How do I change the measurement range? Can I do this using LabView or should something be changed at the DAQ itself?
Karthik
(3,460 Views)
Hi Karthik,
You can configure the channel input limit in LabVIEW using the AI Config VI. The available measurement ranges can be found at the following link:
(3,446 Views)
Sorry, posted too quickly. Here's the link to the specifications for the AT-MIO-16E-2:
href="http://www.ni.com/pdf/products/us/2mhw239-240e.pdf"
(3,442 Views)
You can certainly do filtering in LabVIW but that's not the same as an anti-aliasing filter. No amount of software filtering will rid your signal of aliasing effects.
(3,440 Views)