10-07-2008 07:05 PM
Solved! Go to Solution.
10-09-2008 04:50 AM - edited 10-09-2008 04:51 AM
Hi rsd111,
I understand you are measuring a circuit similar to this one ( http://en.wikipedia.org/wiki/Common_drain )
I get also you're using an external power source with GND connected to Drain and -5V connected to Source with a 100k resistor.
You should be able to measure the Voltage between D and S configuring a differential measurement and connecting Ai0+ to D and Ai0- to S, actually you can youse the Ain you prefer.
You should be able to do the same measurement also in NRSE configuration connecting Ai0 to Drain and AiSense to Source.
The Input impedence of the board is bigger than 10GOhm, it should behave like the oscilloscope in High Impedence DC coupling mode.
6221 Specifications : http://digital.ni.com/manuals.nsf/websearch/8117DF4C5A29C95C862573020061023B
Have a good day.