I am setting up an experiment and I am currently having trouble making the NI 9239 card reading at the correct sample rate.
The detection system is set up as follows:
I already checked if adding or removing the NI 9402 module would make a difference and it does not.
I am using a laser with a 4 kHz repetition rate, and I aim to acquire 4 points for every laser shot, which means that I want to use a sample rate of 16 kHz.
However, when I am making the program run (example attached), it looks like the acquisition is slightly faster than that: indeed if I check what is the rate through the property node "SampClk.Rate" option, it gives me a rate of 16666.7 Hz instead of the wished 16000 Hz.
Digging inside the timing property node I realized that the actual rate is not the one I give to the sample clock as an input, that rate will be rescaled based on the timebase rate and the MaxRate of the device.
Is it somehow possible to get a sampling rate of exactly 16 kHz or exactly another number that is not related to the SampClk.Rate I am seeing in the VI I attached?
I am sorry if the problem might be trivial but I am not an electrical engineer so I am new in this field!
Thanks for help
E
Solved! Go to Solution.
01-25-2024 07:01 AM - edited 01-25-2024 07:07 AM
Sample 13+ points with 50kHz and do a resampling in software?
However since the masterclock and your laser clock are not synchronized, the actual samples will match to the laser pulse +- 1/SR in timing
Any chance to sync the laser pulses timebase with your aquisition timebase?
RTFS:
Mastertimebase is 12.8 MHz
All posible samplerates are 12.8M/256/n = 50k/n (n=1..31)
01-25-2024 09:47 AM
Thanks for answering!
Unfortunately, I can not change the laser pulse timebase...However, I did find a NI9215 card and with this one I can read at sample rate of 16000! I don't understand why I can't do the same with the other one... So I guess I will keep using this one!