Single-ended sensor with external power supply (NI 9201)
Hi, I am hooking up a 0-5 V output accelerometer to a NI 9201 analog voltage input module.
The accelerometer requires between 7 - 36 VDC excitation which I am supplying from a 12V power supply. The power supply is floating - i.e. is not referenced to building ground. It's just a cheap, standard 12V power supply.
Unfortunately, the accelerometer is not behaving as expected - lots of noise and not very responsive, when compared to another accelerometer I have which I know works. I have spent a long time reading up about different ground potentials, single-ended wiring and floating vs ground referenced signals.
As far as I can tell, I am using the correct wiring, as recommended here for a single-ended, floating signal which is not referenced to ground: http://www.ni.com/tutorial/7113/en/
Is there something obvious I'm getting wrong here or is the problem not due to the wiring?
Really appreciate your help with this.
Best wishes,
Robin
Here is the configuration I am using:
(4,002 Views)
Accepted by topic author Robin_Hartley
To check the influence of the powersupply, just replace it with a 9V battery.
To reduce power supply noise :
  • Add a RC filter (R ~10Ohm, C 100nF in parallel with 10µF) in the powersupply, a choke (ferrite to force I+ and I- of the powersupply to be equal) also can help.
  • Use a twisted pair cable from the 9201 and connect the AI-GND/COM at the Sensor (if the sensor has an extra GND connection seperated from power gnd use it!  EDIT: from the picture it doesn't seem so, but if the shield and the GND (black?) wire are connected inside the sensor (check with a DMM), try the shield for power GND and the black for singnal GND.
MEMS come in different tastes 🙂  so maybe the noise it inherent with that sensor.
Using a higher samplerate and applying a lowpass filter and decimation is a common software workaround.
Greetings from Germany
Henrik
LV since v3.1
(3,965 Views)
Hi Henrik - thank you for such an informative and helpful answer.
Just by chance, I brought in a 9V battery today and when I tried the sensor, it worked perfectly! This indicates that the problem is due to noise from the power supply, not a problem with the wiring.
Therefore, the solution is to reduce the power supply noise - the suggestions you have made seem perfect for this so I have accepted your answer as the solution. I will now implement the noise reduction techniques you have suggested.
Thank you kindly for your help - it's very much appreciated 😄
Robin
Oh btw, for future readers, this is a piezoelectric sensor called: 8101-0040X-120 by Measurement Specialities - a good low-cost (~£105) sensor with ~6kHz bandwidth. Enjoy! 🙂
(3,960 Views)
