07-10-2018
03:56 AM
- last edited on
10-21-2025
10:28 AM
by
Content Cleaner
Hi, I am hooking up a 0-5 V output accelerometer to a NI 9201 analog voltage input module.
The accelerometer requires between 7 - 36 VDC excitation which I am supplying from a 12V power supply. The power supply is floating - i.e. is not referenced to building ground. It's just a cheap, standard 12V power supply.
Unfortunately, the accelerometer is not behaving as expected - lots of noise and not very responsive, when compared to another accelerometer I have which I know works. I have spent a long time reading up about different ground potentials, single-ended wiring and floating vs ground referenced signals.
As far as I can tell, I am using the correct wiring, as recommended here for a single-ended, floating signal which is not referenced to ground: https://www.ni.com/en/shop/data-acquisition/sensor-fundamentals/measuring-direct-current-dc-voltage....
Is there something obvious I'm getting wrong here or is the problem not due to the wiring?
Really appreciate your help with this.
Best wishes,
Robin
Here is the configuration I am using:
Solved! Go to Solution.
07-11-2018 03:59 AM - edited 07-11-2018 04:03 AM
To check the influence of the powersupply, just replace it with a 9V battery.
To reduce power supply noise :
MEMS come in different tastes 🙂 so maybe the noise it inherent with that sensor.
Using a higher samplerate and applying a lowpass filter and decimation is a common software workaround.
07-11-2018 04:08 AM - edited 07-11-2018 04:10 AM
Hi Henrik - thank you for such an informative and helpful answer.
Just by chance, I brought in a 9V battery today and when I tried the sensor, it worked perfectly! This indicates that the problem is due to noise from the power supply, not a problem with the wiring.
Therefore, the solution is to reduce the power supply noise - the suggestions you have made seem perfect for this so I have accepted your answer as the solution. I will now implement the noise reduction techniques you have suggested.
Thank you kindly for your help - it's very much appreciated 😄
Robin
Oh btw, for future readers, this is a piezoelectric sensor called: 8101-0040X-120 by Measurement Specialities - a good low-cost (~£105) sensor with ~6kHz bandwidth. Enjoy! 🙂