USB DAQ 6363 measurement bad with several channels
Antonin,
I only use SignalExpress.
If you want, i can send SignalExpress project by mail?
Laurent
Message 11 of 21
(1,854 Views)
Hi Laurent,
Could you send me a schema of your system : current and resitor wiring.
Indeed your resistor has to be between AI- and the - of your voltage source.
Here are two interesting links you could see as well to ensure your wiring :
http://digital.ni.com/public.nsf/allkb/82508CD693197EA68625629700677B70
http://www.ni.com/white-paper/7114/en
Message 12 of 21
(1,851 Views)
Antonin,
One setup example (Agilent multimeter is replaced by DAQ USB-6363) :
Laurent
Message 13 of 21
(1,849 Views)
Laurent,
The NI USB DAQ-6363 is ground-referenced (through your PC), that's why you need to make a conditioning where the shunt resistor will be between the " - " of your power supply and your system.
If the shunt is as on the setup example you gave me, there is a common-mode issue on inputs and a drift in the measurement. To act like your example you would need a NI CompactDAQ configuration with embedded conditioning.
Message 14 of 21
(1,846 Views)
Antonin,
I use DAQ in differential mode, so ground reference is not needed to my mind? Is it false?
Laurent
Message 15 of 21
(1,843 Views)
Hi Laurent,
Even in differential mode the DAQ needs a ground reference.
Let me know,
Cheers
Message 16 of 21
(1,818 Views)
Antonin,
Ok, I will speak with internal HW team to expose connection problem for differential measurement
You think it can be source of my problem (measuremnt diff following sample rate) ?
Regards
Laurent
Message 17 of 21
(1,814 Views)
Antonin,
We are doing power optimisation on embedded system.
We measure current on DCDC converter / LDO input & output by adding serial 0.1ohm shunt. (BGA component)
So we can't use advice from your link (http://digital.ni.com/public.nsf/allkb/82508CD693197EA68625629700677B70)
and use common ground mode
I look at “NI CompactDAQ configuration with embedded conditioning” but don’t find what is different compare to USB-6363 ?
Is there a solution to use our USB-6363 DAQ in proper way?
Laurent
Message 18 of 21
(1,803 Views)
Hi Laurent,
Di you try with the configuration I gave you ? What are the results and do you still have similar issues ?
Let me know please,
Regards
Message 19 of 21
(1,789 Views)
Antonin,
I use this HW setting. I think it fully comply with NI requirements
To simplify i measured directly voltage on shunt that is load (1Ohm/1%)
Generator output is sinusoidal signal
Measurements:
sample to read=200k, Rate=1M, 1 channel
average = 1.01091mV
sample to read=200k, Rate=1M, 2 channels
average = 1.24767mV => ~0.23mV more (+25%) !!!
sample to read=20k, Rate=100k, 2 channels
average = 1.01031mV => need to lower sample rate to have good results
Laurent
Message 20 of 21
(1,783 Views)