Multifunction DAQ
connection of 2 wire transmitters in NI PCI 6221 DAQ card in differential mode.
Hello Everyone
Need to connect 2 wire transmitter in PCI 6221 in differential mode.
Please suggest the wiring details.
(2,382 Views)
AI0 and AI8
68PIN: 68  and  34
37DSub:1  and  20
(2,379 Views)
Thanks for your reply
In a 2 wire transmitter wiring there is close loop power of 24VDC is required for sensor. How we will connect it??
(2,375 Views)
6221 only have power of DC5V
Use another power
If necessary, common-ground:power GND connect to AI GND
(2,363 Views)
The 6221 can only measure voltage natively and not current.
To measure a current transmitter you will need to provide a separate power supply and a load resistor so the 6221 can measure across the load resistor - being careful to ensure the voltages stay within the limits of the 6221.
There are separate modules for reading current directly. This page links to some at the bottom https://www.ni.com/en/shop/data-acquisition/fundamentals--system-design--and-setup-for-the-4-to-20-m...
James Mc
CLA and cRIO Fanatic
My writings on LabVIEW Development are at devs.wiresmithtech.com
(2,361 Views)