04-08-2025 09:20 AM
Dear all,
I am trying to connect a Hamilton EasyFerm PLUS PHI S8 120 pH electrode to a USB-6002 DAQ device using a 1 m pH Cable S7 / Open End cable (also from Hamilton). The electrode is currently connected with the signal wire (+) to AI0 and the reference wire (-) to AIGND with the input being configured as RSE in LabVIEW. While I'm able to calibrate the electrode to give a proper pH measurement for the corresponding buffer solutions, the pH measured for one of my samples is far off the actual pH (pH 7.10 measured with this setup vs. pH 4.8 measured with a calibrated pH meter with a comparable electrode). Furthermore, very long settling times and signal drift can be observed.
Concerning the full setup, a laptop and a LabQ peristaltic pump are also connected to the DAQ device. Currently, the only signal conditioning done to the pH signal is averaging over 10s.
Would adding an op amp or unity gain buffer in line in between the pH electrode and DAQ device help? Or is using a electrode with a built-in pre-amplifier more recommended? I tried to orient myself according to this article:
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019MsvSAE&l=en-CH
04-08-2025 10:05 AM
You need some signal conditioning in between the raw pH electrode and the DAQ.
Adding a unity gain amp (buffer) will help make the measurements better as the electrode output impedance is very high.
04-08-2025 11:30 AM
Thanks for the quick response. Do you have any recommendations concerning components? I'm fairly inexperienced with electronics & circuits so any input will be very helpful.
04-09-2025 09:09 AM
@AtlasCH wrote:
Thanks for the quick response. Do you have any recommendations concerning components? I'm fairly inexperienced with electronics & circuits so any input will be very helpful.
In that case, you can get something like this https://bc-robotics.com/shop/ph-sensor-amplifier/
https://www.amazon.com/Sensor-Electrode-Module-Arduino-Detection/dp/B0DYDBSQNV?gQT=1
If you buy this, you can directly connect it over USB and collect data in LabVIEW.
04-10-2025 01:35 AM
The pH sensor amplifier/module seems like an optimal solution to my problem. I can get the full kit with the sensor to test the viability and then get a S7/BNC cable for my current electrode if it is a viable setup. It also only requires 5 V DC which can be supplied by the USB 6002 itself so the setup will still remain compact. I will definitely try this approach.
Thank you so much for the input, this helped me out a lot.