Dear all

 

I have a problem about LV12 software to config PXI-4070 DMM current function.

 

my vi...NI DMM initial >>>NI DMM reset >>> NI DMM Configure Measurement (Current function) >>> NI DMM Reading >>>>

 

When I testing with my vi, the calibrator (source) error "Current Compliance Voltage Exceeded".

 

but when I testing with DMM front panel 3.0, the calibrator can generated current and PXI-4070 can reading.

 

Please help me to solve this problem.

 

Regard,

Sermsak.

 

 

0 Kudos
Message 1 of 3
(5,926 Views)

The error looks like it is related to a power supply or an SMU that is trying to source something out of its specified range.

 

Info about compliance:

http://zone.ni.com/reference/en-XX/help/370736L-01/ni_dc_power_supplies_help/compliance/

http://zone.ni.com/reference/en-XX/help/370736L-01/nidcpowerviref/nidcpower_query_in_compliance/

 

Things to consider:

1) Where in the code is the error occurring?

2) Can you run example NI-DMM code? (from Help>>find examples>>Hardware I/O>>Modular Instruments>>NI-DMM)

0 Kudos
Message 2 of 3
(5,902 Views)
Solution
Accepted by topic author sermsakk

Dear Sir

I can solve this problem by continues reading.vi.

and the operated current source.

 

thank you for your information.

 

0 Kudos
Message 3 of 3
(5,895 Views)