PXI
NI PXIe 5413 is detected by NI MAX but is not accessible via Instrument Studio
Hello everyone,
I’m currently working with the following setup:
  • Chassis: NI PXIe-1071
  • Waveform generator: NI PXIe-5413
  • Driver board: NI PXIe-8301
  • Digitizer (working fine): NI PXIe-5160
Problem:
The waveform generator is correctly detected in NI MAX, but it does not appear in Instrument Studio. As a result, I cannot use it. In NI MAX, the “Self-Test” option is also unavailable for this device. Opening the NI PXIe-5413 in InstrumentStudio from Device Manager results in the following error (see attached screenshots).
Context:
  • Everything used to work fine.
  • During an automated test sequence I programmed, the PC crashed with a blue screen. Since then, the waveform generator has been unusable.
  • The digitizer still works without issues.
  • I tried moving the entire chassis to another desk, with another PC and another PXIe driver board (as the other PC doesn't have a USB-C port), and in that configuration everything worked properly.
Conclusion so far:
This points to an issue with either the driver board (NI PXIe-8301) or the computer itself.
Any insights, troubleshooting ideas, or suggestions would be greatly appreciated!
Thanks in advance!
(168 Views)
Your PXIe-5413 is not recognized as a FGEN but a RIO device. The device enumeration is not working properly.
Please refer to MXI-Express Compatibility and Connectivity Troubleshooting Guide
https://tmesystems.net/
(130 Views)
Hello ZYOng,
I found a way to resolve my issue. The problem was not with the device being recognized as a "RIO device" but with NI MAX configuration
It seems that it became corrupted during an automated test we conducted.
Thanks to your link I simply reinitialized NI MAX configuration, and the waveform generator works well now. Even tho it still shows up as a "RIO device".