06-10-2008 10:54 AM
06-11-2008 10:43 AM
06-11-2008 02:19 PM
06-12-2008 04:16 PM
06-12-2008 04:32 PM
Hello George!
Actually, when you do a 'disable' using NI-DCPower, the channel 1 output of the PXI-4130 is configured for a 20mA current limit. If you'd like to completely disconnect the SMU from your system, an external relay could be used (you could probably power it using channel 0 of the PXI-4130 if you aren't already using it). Alternatively, you could set a voltage of zero and a small current limit (as low as 4uA) that would limit the flow of this current. In that case you wouldn't actually 'disable' via software.
06-13-2008 07:54 AM