03-26-2007 01:06 PM
03-27-2007 09:41 AM
Hi Richard,
Can you post your calibration report? That error usually occurs when the 5 V reference is out of range, offset enough that Cal Exec cannot adjust it back to place. If this is the case, then the only way to fix it is to send the board in for repair. If this is not the case, we should be able to find out from the report what is causing this.
Hope this helps,
Andrew S.
National Instruments
03-27-2007 11:24 AM
Here is the info from the calibration report. I can email you a PDF of the report and also a screen capture of the error message which states: "Measurement taken during calibration produced an invalid AI gain calibration constant. If performing an external calibration, ensure that the reference voltage passed to the calibration VI or function is correct. Repeat the calibration. If the error persists, conatct National Instruments Technical Support.
CALIBRATION PERFORMANCE TEST DATA
DUT Information
Type: PXI-4461
Tracking Number: 33367
Serial Number: 33367
Notes
Customer Information
Name: Cal Lab
Address:
Purchase Order:
Notes
Environmental Conditions
Temperature: 23.0 C
Humidity: 13.0 %
Operator Information
Operator Name: administrator
Calibration Date: Friday, March 23, 2007
16:27:03
Notes: Error or termination
occurred. This calibration
may not be valid. Error
code: -200718 Error
message: Error -200718
occurred at DAQmx Self
Calibrate.vi at step Self
Calibrate Possible
reason(s): Measurement
taken during calibration
produced an inval
PXI-4461 Serial Number: 33367
Friday, March 23, 2007 16:27:03 Page 1 of 2
Standards used during Calibration
Type Tracking Number Calibration Due Date Notes
Fluke 5500A Multifunction
Calibrator
32261 3/15/2008
DMM 32260 11/21/2007
33250A 29536 10/13/2007
Calibration Results
Test Canceled
Calibration As Found As Left
Test Value Low Limit Reading High Limit PassFail Low Limit Reading High Limit PassFail
N/A N/A N/A N/A N/A N/A N/A N/A N/A
PXI-4461 Serial Number: 33367
Friday, March 23, 2007 16:27:03 Page 2 of 2
03-27-2007 11:28 AM
One more piece of information on this. Thinking that the problem may have been the board itself, we installed a second 4461 in the chassis and tried the calibration procedure. We got the same error which led us to believe that the problem was not in the boards. Both boards came out of function systems.
This is our first attempt at calibrating a PXI card with Calibration Executive.
03-28-2007 05:32 PM
Hey Richard,
I'm checking in with the Calibration team on this to see what they have to say. Just as a check, do you know for a fact that the 5 V reference provided to the 4461 is correct? Also, have you verified your external connections are correct and making contact? I know they're pretty basic suggestions, but you'd be suprised how often that causes a problem.
Let me know if you find anything out, and I'll post back once I hear back from the Cal team.
cheers,
Andrew S.
03-29-2007 07:52 AM
This could be an issue. We did not have any external connections to the board and were waiting for Cal Exec to prompt us with the required connections as it does with other boards.
What should be connected to the module?
Richard
03-29-2007 09:15 AM
Richard,
Let's take a step back. Does this device pass self calibration in MAX? You can check this by selecting the device in MAX, and selecting "Self Calibrate" in the upper right hand corner of MAX - sometimes it's hidden in a drop down menu. If it doesn't pass that, then it will tell us that the driver is not communicating with the HW properly. If it does, we can start looking into connection issues.
Thanks,
Andrew S
03-29-2007 09:33 AM
03-29-2007 03:30 PM - edited 03-29-2007 03:30 PM
Hi Richard,
If this were one card, we would most likley have you send this in for RMA, but seeing as its less likely that you're seeing this on two cards I think we should do more troubleshooting. I would like you to run the attached vi. This reads your internal reverence directly and should give us an idea as to whether your reference is off, or possibly the offset factor is in question. Please try this on both of your 4461's and post the results (Screenshots, saved front panels). I'm looking into seeing how we can calibrate your reference manually, seeing as you have the correct HW but cal exec is failing this first self cal before it can go in and fix the references.
cheers,
Andrew S
Message Edited by stilly32 on 03-29-2007 03:52 PM
03-29-2007 03:53 PM
Sorry , posting problems. Just try one of those
Andrew S