Hi sujeep,
From what I understand you are not getting an accurate reading by measuring with a voltmeter at the SCC-68 terminals or in the Measurement & Automation Explorer (MAX). If this is not the case, please post back with a correction. I think this could be a power or grounding problem.
What do you mean by “to activate sensors in our PCB, we made use of digital signals”? Do you mean that you are using the voltage from the digital signals as an excitation voltage for you sensors? If that is the case, then I recommend an alternative means of excitation. The output of a digital line will not necessarily be exactly at 5 V. Thus, if you are using a digital voltage as an excitation, you will not have reliable excitation.
From what I understand, your sensors are floating signal sources. When measuring a floating signal, we have to ground the signal appropriately. Table 1 found at the bottom of this page shows how to properly ground a floating signal source. I would recommend, however, that you change your measurement type to Referenced Single Ended (RSE) (called Single-Ended Ground Reference in Table 1) since it is the easiest to configure. I recommend testing this configuration on a single sensor before wiring your entire system.
Let me know if this helps. If not, please post back with any
information that is missing from above and additionally: