04-16-2007 01:51 PM - edited 04-16-2007 01:51 PM
I have the Compaq DAQ with the NI 9211 & NI 9205 modules. I’ve written a program that acquires 1 thermocouple reading, 2 voltage readings from load cells and 1 mV reading from an LVDT.
Each of these signals is being displayed on the front panel using the waveform graph VI (See Figure 1). I am taking the 1 output wire from the DAQ assistant and using the split signal VI to display each channel signal individually into waveform graphs (See Figure 2). I have set up custom scales for each channel with no offset.
The problem is that the signals are not static.
For Example, with a constant load of 10 lbs on the load cells, the graphs will indicate 10 lbs then fluctuate + or – 2 lbs. With the LVDT, it will be at position 0.0000” but will increase up to 0.0050” and anything in between. Same story with the thermocouple – constant fluctuation at a constant temperature.
They only way I can get the front panel to display a static value is using the mean VI and numeric indicators. See Figures 3 & 4.
Since I’ve used the mean VI to “average” the signals, I cannot get them to display in the waveform graph. In other words, I have inserted a mean VI between the signal wire of the Split Signal VI (one for each channel) and the individual Waveform graphs. The only way I have succeeded in displaying the “averaged” reading is to use a numeric indicator in the place of the waveform graph.
My concern is that the mean VI is still outputting false readings since it will output a 3 lbs load seeing as it averages between the 0 and 6 lbs load from the bouncing signal.
I have 3 questions:
1. How can I get rid and/or prevent these bouncing signals without the use of the mean VI?
2. Why can I not see a waveform in the graphs once I’ve used the mean VI?
3. How do I get the graph to display a nice horizontal line that moves along the X axisas the data changes? As of right now, the graph is static and only displays the current values.
Figure 1
Figure 3
Figure 4
Thanks
jean-pierre.bedard@nrc-cnrc.gc.ca
Message Edited by jpbedard on 04-16-2007 01:52 PM
04-17-2007 08:47 AM
04-17-2007 10:39 AM
The Bale Load and Friction Force are both load cells with a 0-100 lbs range. They are being exicted externally by a 10VDC Sola Power supply. Both load cells are wired into the 9205 Module. Although not show in my block diagram, all 4 signal wires are being used with the mean VI.
Load Cell Details:
Sensitivity: 3.130mV/V
Input Resistance: 380 Ohms
Output Resistance: 350 Ohms
The LVDT is being powered by a 5VDC sola power supply and is wired into the 9205 module.
The type T thermocouple is wired into the 9211 module. The thermocouple is boucing about + or - 1 degree C. I understand that this is most likely in tolerance and for that reason, I used the mean.VI I simply cannot get it to display on the graph after its been "averaged" let alone get it to display dynamically.
Input Ranges:
Thermocouple:
Max: 100
Min: 0
Scaled Units : deg C
Type T
CJC - Built in
Load Cells:
Max: 100 The min and max are the values expected after custom scaling.
Min: 0
Scaled Units : lbs
Terminal Configuration : Differential
Custom Scaling in MAX
Pre-scaled
Max:0.0313V
Min:0V
Scaled
Max:100lbs
Min:0lbs
LVDT:
Max: 0.092 " The min and max are the values expected after custom scaling.
Min: 0
Scaled Units : inches
Terminal Configuration : Differential
Custom Scaling in MAX
Pre-scaled
Max:-0.301V
Min:-2.813V
Scaled
Max: 0"
Min:0.092"
Hopefully this answers all your questions,
Thanks for your help,
JP