03-06-2006 04:48 PM
03-07-2006 03:46 PM
Hi BadgerNuke,
This is some strange behavior that you are witnessing. At this point, I am not sure what is causing the discrepancy in your data. There are a couple of things that I want you to try that will help us figure out what is causing this problem:
- read the thermocouples as voltages and as temperates in Measurement and Automation Explorer (MAX) to see if this behavior is repeated
- use a shipping example to acquire the data such as the Acquire Thermocouple Samples which is located in the Example Finder in LabVIEW.
Also, would it be possible for you to post some of your data, preferably some of the accurate readings as well as the high temperature readings that are causing the problem? Finally, I am including the link to a tutorial with more information about taking thermocouple readings with the SCXI 1102: How to Use Thermocouples with an SCXI-1102 Module.
Regards,
Hal L.
03-08-2006 09:18 AM
03-08-2006 03:00 PM
03-08-2006 03:04 PM
03-09-2006 05:14 PM
Hi Badgernuke,
There is one more thing that I want you to try. Could you connect the thermocouple to two channels? From there, acquire one channel as a temperature and the other as a voltage and write the data to a file? Could you pinpoint exactly where you start to experience problems in this data? I would like to see how the voltage data varies with the temperature data at extremely high temperatures.
Regards,
Hal L.
03-13-2006 02:25 PM
03-14-2006 06:56 PM
Hi BadgerNuke,
I took a look at your data and your description in your first post was extremely accurate. It looks like some overvoltage protection is being applied when the temperature readings reach around 600ºC.
What is the tip of your thermocouple connected to? It is possible that you could be experiencing a very high common mode voltage. Would it be possible for you to isolate the thermocouples?
Regards,
Hal L.
03-20-2006 09:30 AM