I have four SCC-SG24 Full Bridge Strain Gage modules which I can install in (any) one analog socket of a SG-2345 system. I stick to the J1 socket to rule out it's influence. Two of these modules appear completely dead (both the X and X+8 bridge outputs), one appears to have low sensitivity, and one appears fine (both the X and X+8 bridge outputs). I've installed into each a pair of 30k resistors into R1 and R5 so that I can null the bridge using the trim pots.
I've picked one socket of the SG-2345 and individually connect each of the modules which I've attached to a load cell which contains a full bridge of strain gages. I use the test panel of the NI explorer to read the X or X+8 input to the module and relay in/out the cal resistor And one by one it appears that the X and X+8 input channels are dead on two of these modules except one which has both input channels working fine. I have three tests that I try (1) null the bridge by turning the trim pot, (2) connect the 301k SCC-SG11 shunt resistor to unbalance the bridge, and (3) apply force to the load cell to unbalance the bridge. I carry out these three tests on the X and X+8 channels of the module.
Test (1) Test (2) Test (3)
=======================================================================================================
Module1 X, X+8 ok X, X+8 ok X, X+8 ok
Module 2 X, X+8 no change X, X+8 no change X, X+8 no change
Module 3 X, X+8 no change X, X+8 no change X, X+8 no change
Module 4 X, X+8 no change X, X+8 no change X, X+8 small observable change but
not as much as for Module 1
Two of the dead modules are brand new and the other two had been fiddled around with as we were/are learning to use Labview and the SG-2345, etc.
I'm either destroying these modules or my setup is flawed. How do I trouble shoot my setup?