I have to measure from eight strain gage channels disposed at 100 m from the PC. The cheapest solution I found is to use a SC-2043-SG in a CA-1000 enclosure, a NI USB-6008 and a USB extender.
Question 1: in its User Manual, is mentioned that the outputs of SC-2043-SG are NRSE signals. Does this comply with the input specifications of NI USB-6008 ?
A: the 6008 only has differential and RSE input configurations. However, you can attempt to make it work by using the 6008 in differential mode and connecting AISENSE to the negative pin on each of the channels. NRSE is sort of like differential mode.
Question 2: is there enough space for inserting the NI USB-6008 in the CA-1000 enclosure (using some "home made" fixtures) and are there any thermal problems ?
A: there appear to be enough space to put both in the enclosure. The problem maybe be the thickness of the enclosure. It is 4.39 while the thickness of the 2043 is 4.6. The difference is in the millimeters, but could be a problem. Other than that, there should not be any other problem.
You can check these dimensions in the following PDFs:
http://www.ni.com/pdf/manuals/321738a.pdf
http://www.ni.com/pdf/manuals/371303a.pdf
http://www.ni.com/pdf/manuals/371201a.pdf
Good luck on your project